PLUMBER RICK
05-16-2007, 02:10 AM
i have a situation at a surgical center that is a catch 22:confused:
the center purchased a rebuilt 1200# sterilizer that uses a steam generator that sterilizes the surgical instruments. the problem is the rep had "his people" connect it to the existing location of the older smaller unit. the steam generator has a pressure booster built in to keep up with demand. the actual sterilizer uses another source of water to create a vacuum in the sterilizer chamber. it calls for 8-10 gallons per minute at 50psi residual pressure. the 3 story large medical complex has a 3'' water main with 90psi inlet side and 60psi on the outlet side of the regulator station. seems low as the total height is approx. 40' for the top suites. the sterilizer is on the ground floor, 60psi static. the water feed is only a 1/2'' line with a reduced pressure back flow preventer feeding the unit.
since i can't run this machine to actually get a flow test and a static test without the rep being there, how would i figure the max pressure and volume without running the machine? i do own a real nice digital flow meter and pressure gage. i guess i can disconnect the 1/2'' line and divert it into the floor sink. i also think that the building pressure is set way to low.
is 8-10 gpm with 1/2'' pipe through a rp valve possible with only 60 psi and a drop of 10psi max? the water is used to pull a vacuum using a venturi principal. the machine is only pulling 19'' of mercury and needs 25'' of mercury to operate.
i know there are a lot of variables, but is it just asking too much with these set numbers. if the building ups the regulators to 70psi, can i get 8-10 gallons in your opinion.
i have not tried any real test yet. just a static pressure reading.
the rep who sold this machine to the center, didn't take the water requirements into account. they already spent good money on a 3 phase 50 amp power supply to run the steam generator.
i know there are a lot of unknowns, but the doctor is on the verge of sending it all back. i told him i will see what i can do without the help of the building issues or rep.
so here are the questions:
is 8 to 10 gallons possible at 60psi with a drop of only 10psi and residual of 50psi.
will upping the regulators to 70psi fix the problem? this will affect the entire 3 story complex and i don't know what the extra pressure can affect in the rest of the suites and bathrooms.
or is running a larger 3/4'' line with a new rp valve the only solution?
what would you do?
thanks for your suggestions.
let the debate start:D
rick.
the center purchased a rebuilt 1200# sterilizer that uses a steam generator that sterilizes the surgical instruments. the problem is the rep had "his people" connect it to the existing location of the older smaller unit. the steam generator has a pressure booster built in to keep up with demand. the actual sterilizer uses another source of water to create a vacuum in the sterilizer chamber. it calls for 8-10 gallons per minute at 50psi residual pressure. the 3 story large medical complex has a 3'' water main with 90psi inlet side and 60psi on the outlet side of the regulator station. seems low as the total height is approx. 40' for the top suites. the sterilizer is on the ground floor, 60psi static. the water feed is only a 1/2'' line with a reduced pressure back flow preventer feeding the unit.
since i can't run this machine to actually get a flow test and a static test without the rep being there, how would i figure the max pressure and volume without running the machine? i do own a real nice digital flow meter and pressure gage. i guess i can disconnect the 1/2'' line and divert it into the floor sink. i also think that the building pressure is set way to low.
is 8-10 gpm with 1/2'' pipe through a rp valve possible with only 60 psi and a drop of 10psi max? the water is used to pull a vacuum using a venturi principal. the machine is only pulling 19'' of mercury and needs 25'' of mercury to operate.
i know there are a lot of variables, but is it just asking too much with these set numbers. if the building ups the regulators to 70psi, can i get 8-10 gallons in your opinion.
i have not tried any real test yet. just a static pressure reading.
the rep who sold this machine to the center, didn't take the water requirements into account. they already spent good money on a 3 phase 50 amp power supply to run the steam generator.
i know there are a lot of unknowns, but the doctor is on the verge of sending it all back. i told him i will see what i can do without the help of the building issues or rep.
so here are the questions:
is 8 to 10 gallons possible at 60psi with a drop of only 10psi and residual of 50psi.
will upping the regulators to 70psi fix the problem? this will affect the entire 3 story complex and i don't know what the extra pressure can affect in the rest of the suites and bathrooms.
or is running a larger 3/4'' line with a new rp valve the only solution?
what would you do?
thanks for your suggestions.
let the debate start:D
rick.